Highlights from this day
  • area_percent — coming back to this is uv at 214 or 280 better for peptides 17:18
  • tarpit_tam — how much sample does a lab actually need for identity as well as purity 18:51
  • travel_cooler — tirzepatide is about 4813.5 Da, so 2+ is near 2407.8 and 3+ near 1605.5 19:03
AP

oxidation is about plus 15.99 Da and is usually methionine or tryptophan

coming back to this is uv at 214 or 280 better for peptides

⚠️11📉11🔥3
FI

you need more sample for identity than for purity. plan for that when you post it

AN

shallow gradient

peptidemeter-summary-h-2814.pdf
2 pages · 580 KB · not retained in the public archive
AP

a dimer usually shows up late and at roughly double the mass. the chromatogram alone will not tell you

TC

relative retention time against a standard is comparable between labs. absolute minutes are not

AN

unrelated but a certificate with a number and no chromatogram is asking you to trust the integration you cannot see

TT

how much sample does a lab actually need for identity as well as purity

🤝6
AP

is there any point running a blank if the sample is the question

TC

semaglutide is about 4113.6 Da. [M+2H]2+ lands near 2057.8, [M+3H]3+ near 1372.2

the first peak is the solvent. it is always the solvent

FI

the honest position is that hobby-level interpretation of a chromatogram is worth something but not much

plus 16 is oxidation

TC

214nm sees the amide backbone so it sees everything. 280 only sees aromatics, we shall see